Quantification of pancreas fat by MRS and FATSAT with 3.0T MRI.

Yuan F1, Song B, Tang HH, Chen GW, Li J, Xie Y.

Author information

  • 1Department of Radiology, West China Hospital, Sichuan University, Chengdu, China.

Abstract

OBJECTIVE:

To test the values of MR spectroscopy MRS and fat saturation FATSAT in quantifying fat content in pancreas with 3.0T MR.

METHODS:

Thirty-five healthy volunteers were examined with 3 different MR techniques IP/OP, MRS, and FATSAT, to quantify pancreas fat content FC. The IP/OP result served as a reference standard.

RESULTS:

The IP/OP, FATSAT and MRS detected 8.2 +/- 2.0%, 11.8 +/- 4.7% and 8.8 +/- 4.6% of pancreas fat content, respectively. The IP/OP result correlated with the MRS result r = 0.809, P < 0.001, but not with the FATSAT result r = 0.311, P = 0.069.

CONCLUSION:

For 3.0T MRI, MRS technique can quantify pancreas fat content, but FATSAT can not.

 

Sichuan Da Xue Xue Bao Yi Xue Ban. 2010 Jan;411:145-7, 157.

PMID: 20369491 [PubMed - indexed for MEDLINE]